Calibrating IRZ downhole sensor pressure

Technical forum on ESP sensors. Discuss workshop bench testing, pressure and temperature calibration, vibration checks, and sensor-to-motor assembly
oleg309@yandex.ru
User avatar
Posts: 5
Joined: Thu Nov 18, 2021 6:32 am

Calibrating IRZ downhole sensor pressure

Post by oleg309@yandex.ru »

Calibrating IRZ downhole sensor pressure.

Friends! We stripped an IRZ downhole unit (TMS), found the pressure sensor was bad.

We replaced it, but at 50 atm reference it shows 57, and at 100 atm reference it shows 119. Probably calibration is needed.

Who has run into this? How do you do it?

In the Modbus description I found four addresses "Calibration information," but I read zeros from them, and besides that they are read-only as I understood. Surely the correction coefficients have to be written somewhere...
Original topic
Друзья! Разобрали погружной блок ИРЗ, выяснили, что неисправен датчик давления.

Заменили, но при 50 атм. эталонного, показывает 57, а при 100 атм. эталонного показывает 119. Наверное нужна калибровка.

Кто сталкивался? Как это сделать?

В описании Модбас я нашел четыре адреса "Калибровочная информация", но я считываю с них нули((, и к тому же это только для чтения как я понял. Наверняка куда-то надо записать поправочные коэффициенты...
TMSnick70
User avatar
Posts: 73
Joined: Wed Jan 02, 2019 11:41 am

Re: Calibrating IRZ downhole sensor pressure

Post by TMSnick70 »

This is not a simple job... All downhole units (TMS) are trimmed in software.

I.e. the calibration coefficients are burned into the board and you will not get at them without access to the project. Hardware (resistor) trim was used on the first IRZ units on Microchip controllers, but I have not seen those in a hundred years.

In your case you can pick a resistor (50-200 kOhm) between the arms and that way trim pressure at the top. But that only solves part of the problem.

The second headache is trim at zero pressure. The third is temperature correction.

That is always done for the set - pressure sensor, board.
Original reply
Не простое это дело... Все погружные блоки подстраиваются программно.

Т.е. калибровочные коэффициенты вшиты в плату и добраться до них, не имея доступа проекту, у вас не получится. Аппаратная(резисторная) подгонка применялась на первых блоках ИРЗ на микрочиповских контроллерах, но я их лет сто как не видел.

В вашем случае можно подобрать резистор(50-200кОм) между плечами и таким образом подогнать давление наверху. Но этим можно решить только часть проблемы.

Второй геморой - это подгонка при нуле давления. Третий температурная коррекция.

Это всё делается всегда для комплекта - датчик давления, плата.
oleg309@yandex.ru
User avatar
Posts: 5
Joined: Thu Nov 18, 2021 6:32 am

Re: Calibrating IRZ downhole sensor pressure

Post by oleg309@yandex.ru »

Then why is "Calibration information" (read) present in their Modbus address map? Surely there are addresses for calibration?
Original reply
Для чего у них тогда в карте адресов Модбас присутствует "Калибровочная информация" (для чтения)? Наверняка есть адреса для калибровки?
TMSnick70
User avatar
Posts: 73
Joined: Wed Jan 02, 2019 11:41 am

Re: Calibrating IRZ downhole sensor pressure

Post by TMSnick70 »

I do not know. But definitely not for calibrating the sensor. Calibration information on the BP-103 is never brought out to the SIU (surface interface unit, TMS).
Original reply
Не знаю. Но точно не для калибровки датчика. В наземник никогда не выводится информация по калибровке БП.
oleg309@yandex.ru
User avatar
Posts: 5
Joined: Thu Nov 18, 2021 6:32 am

Re: Calibrating IRZ downhole sensor pressure

Post by oleg309@yandex.ru »

This is from the BP-103 exactly
Original reply
Это из БП как раз
TMSnick70
User avatar
Posts: 73
Joined: Wed Jan 02, 2019 11:41 am

Re: Calibrating IRZ downhole sensor pressure

Post by TMSnick70 »

Very interesting. If you take the nameplate data on any BP-103, pressure accuracy is +-0.5-1%.

For IRZ that is +-2-4 bar. I.e. any BP-103 should hold nameplate accuracy in those limits.

I allow in principle that you can enter BP-103 calibration in the SIU (surface interface unit, TMS), because I have seen that option on American sensors. But I have not seen it on Russian SIUs.

Write the SIU model; if I find this junk on my PC I will look. In any case, if there is calibration, it is meant for trim within nameplate values.

And no more than that.
Original reply
Очень интересно. Если взять паспортные данные на любой БП то там точность по давлению +-0,5-1%.

Для ИРЗ это +-2-4бар. То есть любой БП должен обеспечивать паспортную точность в этих пределах.

Допускаю в принципе что в неземнике можно ввести калибровку БП, потому-что видел такой вариант на американских датчиках. Но на Российских наземниках не видел.

Напишите модель наземника, если найду это гуано у себя на компе - посмотрю. В любом случае, если есть калибровка, то она предназначена для подгонки в пределах паспортных значений.

И не более того.
oleg309@yandex.ru
User avatar
Posts: 5
Joined: Thu Nov 18, 2021 6:32 am

Re: Calibrating IRZ downhole sensor pressure

Post by oleg309@yandex.ru »

SIU (surface interface unit, TMS) TMS-E5
Original reply
Наземный блок ТМС-Э5